X-Ray Forum for NDT Users
It is currently 21.11.2024, 07:41

All times are UTC + 1 hour




Post new topic Reply to topic  [ 1 post ] 
Author Message
PostPosted: 30.12.2019, 18:25 
Offline
User avatar

Joined: 19.10.2019, 15:47
Posts: 137
Similar to the non uniform focal spot size in the beam the intensity distribution is also not uniform in the beam.
Most dose is near the center of the beam and to the edges the dose is reduced. Here is an example of a 20° tube (uncorrected gain image of a detector):

This effect could be reduced by the distance between tube and detector - but for the price of a loss in dose (double distance == quarter of dose).
Using a DDA this intensity distribution shall be corrected by the gain calibration (beside the pixel non uniformity). But also then there is a drawback: Due to the lower dose in the corners the SNR is reduced - in the image of the example by factor of 1.5. If a hole type IQI would be in the center with an CNR of 3.0 and another one near the corner, this one would reach a CNR of 2.0 only - which is less than the mostly required CNR of 2.5.


Top
 Profile  
Reply with quote  
Display posts from previous:  Sort by  
Post new topic Reply to topic  [ 1 post ] 

All times are UTC + 1 hour


Who is online

No registered users


Similar topics

Non uniform size of the focal spot in the beam
Forum: Standards for Focal Spot Measurement for NDT
Author: Klaus
Replies: 0

You cannot post new topics in this forum
You cannot reply to topics in this forum
You cannot edit your posts in this forum
You cannot delete your posts in this forum

Search for:
Jump to:  
cron



Hosted by iphpbb3.com

Impressum | Datenschutz