The title of IEC 60336 is
Medical electrical equipment – X-ray tube assemblies for medical diagnosis – Characteristics of focal spots

The measurement method is similar to EN12543-2; beside the pinhole diaphragm a slit diaphragm may be used as the focal spot length is evaluated only.
Another (important) difference is the threshold of 15% (instead of 10% in EN) of the isodose contour line - with is a simple threshold in IEC 60366 as there is one dimension only.
(taken from 60336 (c) IEC:2005)

And there is a second reason that focal spots are nominated much smaller when measured with IEC60366 - for a nominal focal spot value the dimension may be significant larger. Here is as example the value for a 1mm nominal focal spot value

where the length is double the value of the nominal focal spot value :O

The IEC 60366 is not really relevant for NDT applications - but you should look how the focal spot is measured when you buy a system and calculate the required system-resolution. More information about the influence of the focal spot size in the image could be found at .


Top
 Profile  
Reply with quote  
Display posts from previous:  Sort by  
Post new topic Reply to topic  [ 1 post ] 

All times are UTC + 1 hour [ DST ]


Who is online

No registered users


You cannot post new topics in this forum
You cannot reply to topics in this forum
You cannot edit your posts in this forum
You cannot delete your posts in this forum

Search for:
Jump to:  
cron



Hosted by iphpbb3.com

Impressum | Datenschutz