Re: Main Image Quality Parameters in an X-Ray Image
19.11.2019, 18:37
As the previous posting is too simple, here is a more complete version. In BOLD are the Image Quality Parameter and in the next line the measurement method and IQI.
9. Efficiency (depended to energy (keV)) software tools
There are a few standards for the proper IQIs. In Europe and ISO most of the IQIs are defined in ISO19232 ISO19232-1 Wire (contrast and spatial resolution) ISO19232-2 Step Hole (contrast and spatial resolution) ISO19232-5 Duplex Wire gauge (spatial resolution) - identical to ASTM E2002 In ASTM E07 we have E1025 and in annex of E1742 holetype penetrameter (contrast and spatial resolution) E1647 Plates with recesses (contrast resolution) E2002 Duplex Wire Gauge ((spatial resolution) - identical to ISO19232-5 E747 Wire sets (contrast and spatial resolution) - similar to ISO19232-1 E746 Multi Hole Plates (contrast and spatial resolution) E2737 Duplex Plate with E1025 and E2002 IQIs E2737 Five Groove Wedge (continous wedge with five grooves) E2445 CR Phantom (plate with ten differents IQIs or gauges) - identical to ISO 16371-1 or EN 14784-1