How to prove that your DDA system is still working as required? Discussion about the requirements
Write comments

Inquiry on Section 7.2 of ASTM E2737

06.04.2025, 15:53

Dear Kalus,

I hope this message finds you well.

I am currently working on the implementation of the ASTM E2737-2023 standard and have encountered a question regarding Section 7.2 "Procedure for Measurement of the Offset Level" that I hope you might kindly clarify.

In this section, it is stated that:

Before measurement of the Offset Level, the DDA should be powered-on and not exposed for approximately ten minutes. One image with 30 s acquisition time (for example 1 s frames and averaging all 30 frames) shall be captured without radiation (Offset Image).


I would like to confirm whether the 30-second acquisition time is a strict requirement, or if it is acceptable to deviate slightly from this duration. For example, would it be compliant to acquire 32 frames at 1 frame per second (totaling 32 seconds), or must the total acquisition time be exactly 30 seconds?

Your guidance would be greatly appreciated.

Best regards,
Colin
Write comments



Hosted by iphpbb3.com

Impressum | Datenschutz